Something that has been bugging me for the last runs I did : I use a Powerpak 300 from BioRad in clamped voltage mode. However, the current intensity fluctuates quickly (at 30V, from 0.7A to 1.2A). This is a problem because if the current changes too quickly, the power generator shuts down thinking the circuit is shorted (I have the error message : "change in resistance detected"). If it happens during the night, the system is then not running for a long time, because it has to be manually reset.
Does anyone have the same problem, or is there something wrong with my setup?
0 ·
Comments
Edit: Nicolas what temperature do you run at? My experience is that temperature of the solution influences the specific resistivity of the solution to a great degree. Running a chamber (not to Stanford spec) at 37 vs 50 C has produced an exceedingly wide range of amperage. (increasing temp from 37 to 50 results in a 4x increase of amperage). I will post my calculations.
I do have a previous post which is somewhat relevant to this issue, if you are interested;
http://forum.claritytechniques.org/discussion/6/chamber-current-and-voltage-as-they-relate-to-sample-clearing-and-degradation#latest
I have also noticed that at that temperature and working under a hood my buffer volume decreases (evaporation). This leads to increased concentrations of ions and SDS. Higher ionic concentration affects amperage.
What I do is to have a fill mark and refill constantly with H20 my circulator. (If you fill in Clearing solution you will further increase ion concentration).
This way my amperage remains constant